  619   21 Aug 2009 Exaos LeeForumLink error of "mcnaf"
The "utils/mcnaf.c" uses "camop()",
180:         printf("camop\n");
181:         camop();
But "drivers/camac/camacrpc.c" provides "cam_op()":
void cam_op()
{
}

If you compile each source into an object, you may encounter a link error as
mcnaf.c:(.text+0x3b1): undefined reference to `camop'
collect2: ld returned 1 exit status
The "mcstd.h" provides "camop". So, we need to change "camacrpc.c" to match the definition and fix the link error.
  618   18 Aug 2009 Denis CalvetSuggestionCould not create strings other than 32 characters with odbedit -c "..." command
Hi,
I am writing shell scripts to create some tree structure in an ODB. When 
creating an array of strings, the default length of each string element is 32 
characters. If odbedit is used interactively to create the array of strings, 
the user is prompted to enter a different length if desired. But if the 
command odbedit is called from a shell script, I did not succeed in passing 
the argument to get a different length.
I tried:
odbedit -c "create STRING Test[8][40]"
Or:
odbedit -c "create STRING Test[8] 40"
Or:
odbedit -c "create STRING Test[8] \n 40"
etc. all produce an array of 8 strings with 32 characters each.
I haven't tried all possible syntaxes, but I suspect the length argument is 
dropped. If it has not been fixed in a later release than the one I am using, 
could this problem be looked at?
Thanks,
Denis.
  
  617   10 Aug 2009 Konstantin OlchanskiInfomisc changes from PIENU and T2K
> FYI - committed the last changes from TRIUMF DAQ systems for PIENU and T2K/ND280 FGD and TPC 
> tests:
> svn rev 4541

Also:

- add traps to event buffer code to catch event buffer (shared memory) corruption observed in PIENU
- dynamically allocate some RPC network data buffers to permit better communication between MIDAS clients built with different values of 
MAX_EVENT_SIZE (in T2K/ND280 the default 4 Mbytes is too small for some users, while other users use the default size - this change permits all 
these programs to talk to each other).

K.O.
  616   10 Aug 2009 Konstantin OlchanskiInfomisc changes from PIENU and T2K
FYI - committed the last changes from TRIUMF DAQ systems for PIENU and T2K/ND280 FGD and TPC 
tests:

- mhttpd: add <odb xxx format="%d">xxx</odb>, similar to AJAX ODBget() method
- alarm.c: if alarm stops the run, log a message (sometimes it is hard to tell "why did this run stop?!?") 
use DETACH transition (was ASYNC - does not follow requested transition sequencing, now calls 
mtransition helper). Also verified that alarm handler always runs on the main computer - for remote 
clients, alarms are processed inside the corresponding mserver process.
- midas.c: event buffer fixes:
-- mserver 100% cpu busy loop if event buffer is full
-- consolidate event buffer cleanup into one routine. do things similar to odb cleanup - check for client 
pid, etc.
-- do not kill clients that have the watchdog timeout set to zero.

svn rev 4541
K.O.
  615   04 Aug 2009 Exaos LeeForumThe contents of the attachment
As requested from K.O., I paste the "00README.txt" as the following:
#-*- mode: outline -*-
#-*- encoding: utf-8 -*-
#AUTHOR: Exaos Lee <Exaos DOT Lee AT gmail DOT com>

* Directories
  +--> 00README.txt : This file
  |
  +--> bustester : Directory contains utilities for VME bus testing
  |
  +--> modules   : APIs to handle VME modules
  |
  +--> pyutil    : Uitilies in Python, including PyMVME
  |
  +--> sis3100   : Provide lib_sis3100mvme.a/so using with "mvmestd.h"

* Utilities in Python

** PyMVME module

   The module "PyMVME" provides the following stuff:
      a. class StdVME
      	 -- contains standard VME informations.
      b. class MVME_INTERFACE
      	 -- the C structure MVME_INTERFACE wrapped in Python
      c. dict MVME_STATUS
      	 -- the return information defined in "mvmestd.h"
      d. the related useful aliases from "mvmestd.h"
      	 -- including "mvme_addr_t", "mvme_locaddr_t", "mvme_size_t"
      e. class MvmeDev
      	 -- the major class which provides methods to access VME bus.

   You may find examples of how to use module "PyMVME" from "find_caen.py" or
   scripts in dir "test". All of the examples are using "lib_sis3100mvme.so".
   You may find information later in this introduction.

** find_caen.py

   The script to find VME modules from CAEN. Now, it is still in test status
   and can only find ADCs, TDCs or QDCs.

* SIS3100 library to be used togather with "mvmestd.h"

  The directory "sis3100" contains sources to build libraries as the following:
  a. lib_sis3100.a     -- APIs declared in "sis3100_vme_calls.h"
  b. lib_sis3100mvme.a -- APIs declared in "mvmestd.h". It also contains the
     		       	  same APIs from lib_sis3100.a

  If you want to use shared libraries, especially when you are using utilities
  wrote in Python, you may rebuild the libraries as the following:

    $ cd sis3100
    $ make shared

* APIs to handle VME modules

** vadc_caen.h/c

   Provides APIs to handle ADC-type modules from CAEN, including:
      a. ADCs --- V785, V785N
      b. TDCs --- V775, V775N
      c. QDCs --- V792, V792N

* VME bus testers

  Still under development.


  614   04 Aug 2009 Exaos LeeForumAbout python interface
Coding in Python is faster than in C (but running slower). So, some python interfaces are useful for testing purpose. I hope you may like the PyMVME module for VME bus testing.
  613   04 Aug 2009 Exaos LeeForumScripts to handle MIDAS sessions
Hi, all again

I have some scripts in "bash" and "Python" to handle MIDAS sessions. Please see the attached utils4midas.zip. I didn't write instructions in detail of how to use them. But I think they are very simple. You may find how to use them by reading the codes and example files.

Best wishes.

Exaos Lee
Attachment 1: utils4midas.zip
  612   04 Aug 2009 Exaos LeeForumVME-related codes contribution
Hi, all

I have some codes while using MIDAS. I upload them here. They are tested with
SIS3100. I haven't other VME controllers, so I don't know whether they work with
other controllers. I just hope that they are helpful. You may find information
from the file "00README.txt" in the package vme4midas.zip. My English is limited, I just hope
that you may catch my ideas. Smile

All my best.

Exaos Lee
Attachment 1: vme4midas.zip
  611   03 Aug 2009 Konstantin OlchanskiForumHow to distinguish the status and value returned from "mvme_read_value(...)"
> uint32_t mvme_read_value(MVME_INTERFACE * vme, mvme_addr_t vme_addr);
> Question: How to distinguish the status and value returned?

On VME interfaces using the Universe and tsi148 PCI-VME bridges, your question has no meaning.

The VME address space is directly mapped into the PCI address space, then mmap()ed into your 
program address space. Internally mvme_read_value() is "return *(uint32_t*)(mmap_base + vme_addr);" 
and there is no such thing as "status".

Physically on the VME bus, for single-word VME cycles (mvme_read_value), there are only 2 error 
conditions, an AS or a DS timeout, and these bridges return the bit pattern 0xFFFFFFFF for either error, 
the same as the traditional VME bus always worked (i.e. before PCI, before ISA, back when the VME bus 
*was* the main CPU-memory-IO bus).

So the answer to your question is "yes". If mvme_read_value() returned 0xFFFFFFFF, there was a VME 
bus timeout because the board you are trying to address a) is not installed, b) was unplugged, c) does 
not decode the address you tried to access (maybe you used the wrong AM code or the wrong data 
width).

With Universe and tsi148 PCI-VME bridges, the mvme_read() call runs the DMA interface that can issue 
block transfer cycles on the VME bus. These DMA interfaces have interesting error handling, but 
basically, they only tell you the estimated VME address at which the AS or DS timeout or BERR has 
occurred. For sane VME boards, DMA errors mean very basic breakage of the VME crate and VME board.

With non-directly attached VME interfaces, i.e. the SIS3100, you can also have communication errors. I 
do not know how those are reported by the SIS3100 Linux drivers, and I do not know how the MIDAS 
driver reports them. But I do know that if you see those errors, your interface is very broken and VME 
bus errors are the least of your worries.

P.S. There also exist PCI bus errors, they also return the bit pattern 0xFFFFFFFF and mean basic 
breakage inside your computer. PCI-PCI and PCI-host bridges have special registers you can read to 
find out the exact cause of the error ("your computer is broken").

K.O.
  610   03 Aug 2009 Stefan RittForumHow to distinguish the status and value returned from "mvme_read_value(...)"

Exaos Lee wrote:
The definition of mvme_read_value is as the following:
unsigned int EXPRT mvme_read_value(MVME_INTERFACE * vme, mvme_addr_t vme_addr);

Read single data from VME bus. Useful for register access. See example in
mvme_open()

Parameters:
    	*vme 	  VME structure
    	vme_addr  source address (VME location).

Returns:
    MVME_SUCCESS


Question: How to distinguish the status and value returned? Should the definition be something like
int EXPRT mvme_read_value(MVME_INTERFACE *mvme, mvme_addr_t vme_addr, unsigned int *var);


This function is a shortcut when you want something like
  printf("%d\n", mvme_read_value(...)); 

and you know that the status is ok. Without this function, you would need to define a variable
  unsigned long d;
  mvme_read_value(..., &d);
  printf("%d\n", d);

so the above function is just a handy shortcut. If you want to see the status however, you can call the "normal" function mvme_read:
  status = mvme_read(..., &d, adr, 4); 
  609   03 Aug 2009 Exaos LeeForumHow to distinguish the status and value returned from "mvme_read_value(...)"
The definition of mvme_read_value is as the following:
unsigned int EXPRT mvme_read_value(MVME_INTERFACE * vme, mvme_addr_t vme_addr);

Read single data from VME bus. Useful for register access. See example in
mvme_open()

Parameters:
    	*vme 	  VME structure
    	vme_addr  source address (VME location).

Returns:
    MVME_SUCCESS


Question: How to distinguish the status and value returned? Should the definition be something like
int EXPRT mvme_read_value(MVME_INTERFACE *mvme, mvme_addr_t vme_addr, unsigned int *var);
  608   18 Jul 2009 Exaos LeeForumjorway73a.c, Data taking hangs in the middle of run
> > > Could you give more info on your setup:
> > > - CAMAC controller model
> > Jorway 73A, we have three in hand and the problem doesn't depend on which controller
> > we were using.
> 
> Dawei sent me a copy of his jorway73a.c scsi-camac driver. It is quite different from the
> file in the MIDAS distribution. Dawei tells me that the file from the MIDAS distribution
> does not compile. Stack traces from Dawei indicate a hang in this modified jorway73a.c
> scsi-camac driver.
> 
> K.O.

I encountered too that the jorway73a.c cannot work for my SCM-301 CAMAC driver. The
"jorway73a.c" distributed with MIDAS seems to work with Jorway 73a reversion > 300. But my
module has the reversion number 203. :-( I hope you can paste the modified version here so
that I can try it if I have spare time.
Regards.
  607   08 Jul 2009 Konstantin OlchanskiForumjorway73a.c, Data taking hangs in the middle of run
> > Could you give more info on your setup:
> > - CAMAC controller model
> Jorway 73A, we have three in hand and the problem doesn't depend on which controller
> we were using.

Dawei sent me a copy of his jorway73a.c scsi-camac driver. It is quite different from the
file in the MIDAS distribution. Dawei tells me that the file from the MIDAS distribution
does not compile. Stack traces from Dawei indicate a hang in this modified jorway73a.c
scsi-camac driver.

K.O.
  606   06 Jul 2009 Dawei LiuForumData taking hangs in the middle of run
Hi Pierr-Andre,


> Hi Dawei,
> 
> Could you give more info on your setup:
> - CAMAC controller model

Jorway 73A, we have three in hand and the problem doesn't depend on which controller
we were using.

> - ADC model

LeCroy 2249W. We also tried two other modules LeCroy 2249A. Same problem.

> - LAM setting

The poll and ADC reading codes are basically from Midas distribution.

> - Mode of polling (on module or on CC)

Polling on CC. I also tried to add a timeout code reading ADC, didn't solve the problem.

> - Are you still going through the poll_event() after hang up?

That's I don't know. I believe the problem happens between finishing reading one event
and passing the control back to poll_event.

> - Do you have the same problem at low rate (100Hz)?

The rate we are currently running is about 400 Hz, it has the same problem. We will
try lower rate more.

Thanks,

Dawei

> 
> Pierre-André
> > Hi,
> > 
> > We are using midas to read ADC. It sometimes hung in the middle of data taking.
> > We tried to disable analyzer and only run with frontend. The problem still
> > exists. We tried to use different crate, different CAMAC controller and
> > different ADC module. All these did not solve the problem. We use polled method
> > to read data. We have dataway display unit so we know that it hung always after
> > it executed CAMAC command F9, which is after finishing one data taking and clear
> > the ADC for the next data taking. The data rate is about 1 KHz. It is random for
> > how long it takes for the system to hang.
> > 
> > Any ideas ?
> > 
> > Thanks,
> > 
> > Dawei Liu
  605   03 Jul 2009 Pierre-Andre AmaudruzForumData taking hangs in the middle of run
Hi Dawei,

Could you give more info on your setup:
- CAMAC controller model
- ADC model
- LAM setting
- Mode of polling (on module or on CC)
- Are you still going through the poll_event() after hang up?
- Do you have the same problem at low rate (100Hz)?

Pierre-André
> Hi,
> 
> We are using midas to read ADC. It sometimes hung in the middle of data taking.
> We tried to disable analyzer and only run with frontend. The problem still
> exists. We tried to use different crate, different CAMAC controller and
> different ADC module. All these did not solve the problem. We use polled method
> to read data. We have dataway display unit so we know that it hung always after
> it executed CAMAC command F9, which is after finishing one data taking and clear
> the ADC for the next data taking. The data rate is about 1 KHz. It is random for
> how long it takes for the system to hang.
> 
> Any ideas ?
> 
> Thanks,
> 
> Dawei Liu
  604   02 Jul 2009 Dawei LiuForumData taking hangs in the middle of run
Hi,

We are using midas to read ADC. It sometimes hung in the middle of data taking.
We tried to disable analyzer and only run with frontend. The problem still
exists. We tried to use different crate, different CAMAC controller and
different ADC module. All these did not solve the problem. We use polled method
to read data. We have dataway display unit so we know that it hung always after
it executed CAMAC command F9, which is after finishing one data taking and clear
the ADC for the next data taking. The data rate is about 1 KHz. It is random for
how long it takes for the system to hang.

Any ideas ?

Thanks,

Dawei Liu
  603   26 Jun 2009 Konstantin OlchanskiInfomhttpd now uses mtransition
> > > mhttpd function for starting and stopping runs now uses cm_transition(DETACH) which spawns an 
> > > external helper program called mtransition to handle the transition sequencing.

Problem reported by Stefan - user presses the "stop the run" button, and the web page comes back saying "running" as if the button did not work. This is 
confusing. It happens because mtransition did not start yet - we have a race condition against it.

To improve this situation, mhttpd now remembers that a start/stop button was pushed and displays a message "Run start/stop requested" until it detects 
that mtransition started and set "runinfo/transition in progress" (or the run state changed).

svn rev 4520
K.O.
  602   25 Jun 2009 Stefan RittBug ReportTR_STARTABORT transition, mlogger duplicate event problem
> Are you sure? Only clients that register themselves to receive the TR_STARTABORT transition (via cm_register_transition()) will receive this transition.
> 
> As of now, the only client that registers and receives this transition is mlogger.
> 
> I also confirm that old clients that know nothing about TR_STARTABORT are *not* sent this transition. (this is tested).

Ok, then we are fine.
  601   25 Jun 2009 Konstantin OlchanskiBug ReportTR_STARTABORT transition, mlogger duplicate event problem
> > Stefan suggested implementing a new transition, TR_STARTABORT, issued if TR_START fails. mlogger can use it to cleanup open files, etc, similar to TR_STOP.
> > 
> > This is now implemented. In mlogger, TR_STARTABORT is similar to TR_STOP, but deletes open output files and does not save end-of-run information into databases, etc. mfe.c does not handle this trnasition yet, but I 
> > plan to add it - to fix the observed situations where the run failed to start, but some equipment does not know about it and continues to generate events and send data.
> > 
> > svn rev 4514
> > K.O.
> 
> There is one problem with the TR_STARTABORT: If you combine old and new clients they will crash, since the old clients don't know anything about TR_STARTABORT. The way to prevent this is to increase the Midas version from 
> 2.0.0 to 2.1.0. Then you will get a warning if you mix clients. Please test this and commit the change if it works.

Are you sure? Only clients that register themselves to receive the TR_STARTABORT transition (via cm_register_transition()) will receive this transition.

As of now, the only client that registers and receives this transition is mlogger.

I also confirm that old clients that know nothing about TR_STARTABORT are *not* sent this transition. (this is tested).

K.O.
  600   25 Jun 2009 Stefan RittForumFrontend and manual trigger question
> I would like to add manual trigger capability and so I added the EQ_MANUAL_TRIG
> flag to the "CAEN" equipment type but the problem is that the framework calls
> directly the readout function on "Midas manual trigger". To trigger manually the
> CAEN ADC's I have to write some registers and therefore I either need to have a
> function called before the readout function or be able in the readout function
> to know if the call has been triggered by the poll function or "Midas manual
> trigger". I tried to check the value *((DWORD *)pevent) but it seems to be a
> well defined and meaningful value only when the readout function call is
> triggered by the poll function.

Actually there is no way to figure out if your readout function is called normally or 
manually triggered. So I modified the framework to add this functionality. In your 
readout routing you can how call

  flag = DATA_SIZE(pevent);

If flag is zero, this is a normal call, if it's one, it's a manual trigger. To get 
this functionality, you have to update midas.h and mfe.c from the repository (rev. 
4519).