  175   24 Nov 2004 chris pearsonInfomidas on 64bit opteron
   Midas, version 1.9.5 of 7th October, was installed, with a few changes, on a
64 bit opteron computer, running linux.  For this processor, as for the alpha
processor, long integers and addresses are 64 bits.  We added a new flag in the
Makefile,

250a251
> ARCH   = $(shell uname -m)
377a379,381
> ifeq ($(ARCH),x86_64)
> OSFLAGS := $(OSFLAGS) -DX86_64
> endif

and extended the alpha-specific definitions, of DWORD and PTYPE, in midas.h to
include this case,

549c549
< #ifdef __alpha
---
> #if defined(__alpha) || defined(X86_64)
598c598
< #ifdef __alpha
---
> #if defined(__alpha) || defined(X86_64)

apart from this, there are a large number of cases where pointers are cast to
integers, without using the PTYPE definition.  These all need to be changed by
hand, although these conversions should probably be removed anyway - in almost
all cases they are unnecessary, as just differences are being calculated.

There were also a number of warnings, which we ignored, where printf format
strings specified long integers, but the argument was not a long integer.  Casts
should probably be added in all cases where the type of the argument can vary
depending on the machine.

A midas analyser was made, which was able to successfully replay some data, but
this was all that was tested.

Chris
  176   25 Nov 2004 chris pearsonForumuse of assert in mhttpd
   We've had mhttpd aborting regularly since upgrading from midas-1.9.3.  This
happens during elog queries, and is due to an elog file that was incorrectly
modified by hand.  The modification to the file occurred 6 months ago.
   el_retrieve(midas.c:15683) now has several assert statements, one of which
aborts the program on reading the bad entry.

   Why is assert used, instead of an error return from the function (if
necessary), and maybe an error message in the log file?  Assert statements are
often removed, using NDEBUG, for normal use.

Chris

   The problem elog entry had one character removed, so end-of-file came before
the end of the message.  This could probably occur without the file being
altered, if the disk containing the elog fills.
  730   02 Nov 2010 chris pearsonInfomhttpd: Extra entries on status page
   A couple of experiments at triumf wanted certain important odb variables
displayed on their status page.  (There was already the possibility to show the
run comment)
   A new folder "/Experiment/Status Items" was created containing links to the
variables of interest, these items are show on the status page, under the run
comment (if any), in 3 columns.

the code from mhttpd.c:show_status_page()
between
   /*---- run comment ----*/
and
   /*---- Equipment list ----*/
is attached
  702   12 Jun 2010 hai quForumcrash on start run
Dear experts,

I use fedora 12 and midas 4680. there is problem to start run when the frontend
application runs fine. 


# odbedit -c start


Starting run #18
[midas.c:8423:rpc_client_connect,ERROR] timeout on receive remote computer info: 
[midas.c:3659:cm_transition,ERROR] cannot connect to client
"feTPCPacketReceiver" on host tpcdaq0, port 36663, status 503
[midas.c:8423:rpc_client_connect,ERROR] timeout on receive remote computer info: 
[midas.c:4880:cm_shutdown,ERROR] Cannot connect to client 'frontend' on host
'hostname', port 36663
[midas.c:4883:cm_shutdown,ERROR] Killing and Deleting client
'feTPCPacketReceiver' pid 24516
[midas.c:3857:cm_transition,ERROR] Could not start a run: cm_transition() status
503, message 'Cannot connect to client 'frontend''
Run #18 start aborted
Error: Cannot connect to client 'frontend'

11:03:42 [Logger,INFO] Deleting previous file "/home/daq/Run/online/run00018.mid"

11:03:42 [Logger,INFO] Client 'feTPCPacketReceiver' on buffer 'SYSMSG' removed
by cm_watchdog because client pid 24516 does not exist

11:03:42 [Logger,ERROR] [system.c:563:ss_shm_close,ERROR]
shmctl(shmid=7274511,IPC_RMID) failed, errno 1 (Operation not permitted)

11:03:42 [ODBEdit,INFO] Run #18 start aborted
==========================================================================

there are several ethernet cards on the host machine. eth0 connect the host
machine to the gateway machine and the front end application listen to eth1 for
the incoming data packets:

eth0      Link encap:Ethernet  HWaddr xx:xx:xx:xx:xx:xx  
          inet addr:10.0.1.1  Bcast:10.0.1.63  Mask:255.255.255.0
          inet6 addr: fe80::f6ce:46ff:fe99:709b/64 Scope:Link
          UP BROADCAST RUNNING MULTICAST  MTU:1500  Metric:1
          RX packets:470870 errors:0 dropped:0 overruns:0 frame:0
          TX packets:515987 errors:0 dropped:0 overruns:0 carrier:0
          collisions:0 txqueuelen:1000 
          RX bytes:345000246 (329.0 MiB)  TX bytes:377269124 (359.7 MiB)
          Interrupt:17 

eth1      Link encap:Ethernet  HWaddr xx:xx:xx:xx:xx:xx   
          inet addr:10.0.1.2  Bcast:10.255.255.255  Mask:255.0.0.0
          inet6 addr: fe80::226:55ff:fed6:56a9/64 Scope:Link
          UP BROADCAST RUNNING MULTICAST  MTU:1500  Metric:1
          RX packets:0 errors:0 dropped:0 overruns:0 frame:0
          TX packets:15 errors:0 dropped:0 overruns:0 carrier:0
          collisions:0 txqueuelen:1000 
          RX bytes:0 (0.0 b)  TX bytes:1836 (1.7 KiB)
          Memory:ec180000-ec1a0000 


thanks for hints
  704   14 Jun 2010 hai quForumcrash on start run
> - does your feTCPPacketReceiver die during the start-of-run? Maybe you do some segfault 
> int he begin-of-run routine. Can you STOP a run?
when start a run, it bring the mtransition process and I guess the server try to talk to the
client, then it fails and the frontend application get killed since not response.

>> When you try to stop your fe from 
> odbedit with  # odbedit -c "shutdown feTCPPacketReceiver"

it gets
[midas.c:8423:rpc_client_connect,ERROR] timeout on receive remote computer info: 
[midas.c:4880:cm_shutdown,ERROR] Cannot connect to client
"feTPCPacketReceiver" on host 'tpcdaq0', port 35865
[midas.c:4883:cm_shutdown,ERROR] Killing and Deleting client
'feTPCPacketReceiver' pid 27250
Client feTPCPacketReceiver not active


what does this error mean? :
11:03:42 [Logger,ERROR] [system.c:563:ss_shm_close,ERROR]
shmctl(shmid=7274511,IPC_RMID) failed, errno 1 (Operation not permitted)


thanks
hai

p.s. that code runs fine on my laptop with ubuntu 9, so that also be possible that somewhere
my configuration not right to cause problem
  2330   08 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
Hi all,

I am having some issues getting the ODBINC command to work within the MIDAS 
sequencer. I am trying to increment one of the ODB values but it is returning a 
mismatch data-type size error (see attached image).

All the ODB variables are MIDAS data-type FLOAT and should all be 32-bit values, 
but for some reason MIDAS is thinking they are 4-bit values. I have tried creating 
new ODB keys of type INT32, UINT32 and DOUBLE but they all give the same error.

If anybody has any suggestions I would really appreciate some help:)

Thanks



 
  2332   09 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> Please post the output of odbedit "ls -l" for /eq/ar.../variables. (you posted the 
> variable name as an image, and I cannot cut-and-paste the odb path!). BTW data size 4 is 
> correct, 4 bytes for INT32/UINT32/FLOAT. For DOUBLE it should be 8. For you it prints 32 
> and this is wrong, we need to see the output of "ls -l".
> K.O.

Hi,

Thanks for getting back to me regarding this. The output of "ls -l" is:

[local:mu3eMSci:S]/>cd Equipment/ArduinoTestStation/Variables
[local:mu3eMSci:S]Variables>ls -l
Key name                        Type    #Val  Size  Last Opn Mode Value
---------------------------------------------------------------------------
_T_                             FLOAT   1     4     1h   0   RWD  20.93
_F_                             FLOAT   1     4     1h   0   RWD  12.8
_P_                             FLOAT   1     4     1h   0   RWD  56
_S_                             FLOAT   1     4     1h   0   RWD  5
_H_                             FLOAT   1     4     60h  0   RWD  44.74
_B_                             FLOAT   1     4     60h  0   RWD  18.54
_A_                             FLOAT   1     4     1h   0   RWD  14.41
_RH_                            FLOAT   1     4     1h   0   RWD  41.81
_AT_                            FLOAT   1     4     1h   0   RWD  20.46
SP                              INT16   1     2     1h   0   RWD  10

Many Thanks
Jago
  2338   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> I tried following script:
> 
> ODBSET /Equipment/ArduinoTestStation/Variables/_S_, 10
> 
> LOOP 10
>   WAIT seconds, 3
>   ODBINC /Equipment/ArduinoTestStation/Variables/_S_
> ENDLOOP
> 
> and it worked as expected. So I conclude the problem must be in your script. Probably a typo in 
> the ODB path pointing to a 32-byte string instead to a 4-byte float.
> 
> Stefan  

Hi Stefan,

Cheers for the reply. I believe the syntax we are using is correct. I have tried copying the script 
you used above and it results in the same error as before. Perhaps something is going wrong 
elsewhere - I will take another look today.

Jago
  2339   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > 
> > [local:mu3eMSci:S]/>cd Equipment/ArduinoTestStation/Variables
> > [local:mu3eMSci:S]Variables>ls -l
> > Key name                        Type    #Val  Size  Last Opn Mode Value
> > ---------------------------------------------------------------------------
> > _T_                             FLOAT   1     4     1h   0   RWD  20.93
> > _F_                             FLOAT   1     4     1h   0   RWD  12.8
> > _P_                             FLOAT   1     4     1h   0   RWD  56
> > _S_                             FLOAT   1     4     1h   0   RWD  5
> > _H_                             FLOAT   1     4     60h  0   RWD  44.74
> > _B_                             FLOAT   1     4     60h  0   RWD  18.54
> > _A_                             FLOAT   1     4     1h   0   RWD  14.41
> > _RH_                            FLOAT   1     4     1h   0   RWD  41.81
> > _AT_                            FLOAT   1     4     1h   0   RWD  20.46
> > SP                              INT16   1     2     1h   0   RWD  10
> > 
> 
> This looks okey, so we still have no explanation for your error. Please post your sequencer 
> script?
> 
> K.O.

Hey, thanks for getting back to me

We are fairly confident the syntax is correct. Having tried the test script posted by Stefan:

> ODBSET /Equipment/ArduinoTestStation/Variables/_S_, 10
> 
> LOOP 10
>   WAIT seconds, 3
>   ODBINC 

The same error is returned:/

We will take another look today.

Jago
  2341   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> Just post here a minimal script which produces the error, so that I can try myself.
> 
> ... and make sure that you have the latest develop version of midas.
> 
> Stefan

Here is the simplest script which produces the error:

WAIT seconds, 3
ODBINC /Equipment/ArduinoTestStation/Variables/_S_

I noticed that "Jacob Thorne"  in the forum had the same issue as us in Novemeber last 
year. Indeed we have not installed any later versions of MIDAS since then so we will 
double check we have the latest version.

Jago
  2343   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > I noticed that "Jacob Thorne"  in the forum had the same issue as us in Novemeber last 
> > year. Indeed we have not installed any later versions of MIDAS since then so we will 
> > double check we have the latest version.
> 
> As you see from my reply to Jacob, the bug has been fixed in midas since then, so just 
> update.
> 
> Stefan

We have tried updating using both:

git submodule update --init --recursive

and:

git pull --recurse-submodules

But the error still persists. Is there another way to update which we are missing?

Cheers
Jago
  Draft   15 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > But the error still persists. Is there another way to update which we are missing?
> 
> The bug was definitively fixed in this modification:
> 
> https://bitbucket.org/tmidas/midas/commits/5f33f9f7f21bcaa474455ab72b15abc424bbebf2
> 
> You probably forgot to compile/install correctly after your pull. Of you start "odbedit" and do 
> a "ver" you see which git revision you are currently running. Make sure to get this output:
> 
> MIDAS version:      2.1
> GIT revision:       Fri Feb 11 08:56:02 2022 +0100 - midas-2020-08-a-509-g585faa96 on branch 
> develop
> ODB version:        3
> 
> 
> Stefan

Hey Stefan,

We are running the GIT revision midas-2020-08-a-509-g585faa96:

[local:mu3eMSci:S]/>ver
MIDAS version:      2.1
GIT revision:       Tue Feb 15 16:31:07 2022 +0000 - midas-2020-08-a-521-ge43ea7c5 on branch develop
ODB version:        3


which is still giving the error unfortunately. 
  2346   16 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > But the error still persists. Is there another way to update which we are missing?
> 
> The bug was definitively fixed in this modification:
> 
> https://bitbucket.org/tmidas/midas/commits/5f33f9f7f21bcaa474455ab72b15abc424bbebf2
> 
> You probably forgot to compile/install correctly after your pull. Of you start "odbedit" and do 
> a "ver" you see which git revision you are currently running. Make sure to get this output:
> 
> MIDAS version:      2.1
> GIT revision:       Fri Feb 11 08:56:02 2022 +0100 - midas-2020-08-a-509-g585faa96 on branch 
> develop
> ODB version:        3
> 
> 
> Stefan

We we're having some problems compiling but have got it sorted now - thanks for your help:)

Jago
  2412   20 Jun 2022 jianrunBug ReportError in "midas/src/mana.cxx"
Dear Midas developers,

When we are running the examples in $MIDASSYS/examples/experiment/, we meet some 
problems when analyzing the results:
1. When we analyze the data using the analyzer: ./analyzer -i run00001.mid -o 
run00001.rz  , we find some bugs: 
"
Root server listening on port 9090...
Running analyzer offline. Stop with "!"
[Analyzer,ERROR] [mana.cxx:1832:bor,ERROR] HBOOK support is not compiled in
[Analyzer,INFO] Set run number 6 in ODB
Load ODB from run 6...OK
run00006.mid:2680  events, 0.00s
"
We think this occurs in the "midas/src/mana.cxx ". How can we solve this?

2. When we analyze the above data, an error also occurs: 
[Analyzer,ERROR] [odb.cxx:847:db_validate_name,ERROR] Invalid name 
"/Analyzer/Tests/Always true/Rate [Hz]" passed to db_create_key_wlocked: should 
not contain "["

We simply fixed that just by replacing the "Rate [Hz]" with "Rate" in the 
test_write in midas/src/mana.cxx 
We are curious whether you can fix the problem permanently in the next version, 
or we are not running the code properly. Thanks!
  1668   28 Aug 2019 lcpForumHistory plot problems for frontend with multiple indicies
hi, 

> > That makes things more 
> > complicated than needed. In the normal FE framework, you can define either several equipment 
> > served by one frontend, or even one equipment linked to several devices. In the MEG experiment 
> > we have one slow control frontend controlling ~100 devices without problem. In the old days 
> there 
> > was a problem that some slow devices could throttle the readout, but since the invention of 
> multi-
> > threaded slow control equipment, each device gets its own thread so they don't block each 
> other.
> 

I agree with Stefan, that it's probably better to run a multi-threaded setup, than individual frontends.

The only place I've ever used the frontend index on startup is when I was testing and building
an eventbuilder.

https://midas.triumf.ca/MidasWiki/index.php/Event_Builder#Example

This might explain, why your history is swapping between frontends, as in the event builder, it gets
reconstructed.

Maybe this helps...

LCP


> Perhaps things have changed in the 10 years since the FGD SC code was written. I can do it 
> differently but doing it that way seemed naturual since around 90% of the frontend code that I
> have see does it that way.
  700   08 Jun 2010 nicholasForumcheck out from svn
do: svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/midas/trunk
midas
shows: ssh: connect to host savannah.psi.ch port 22: Connection timed out
svn: Connection closed unexpectedly
  701   08 Jun 2010 nicholasForumcheck out from svn
> do: svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/midas/trunk
> midas
> shows: ssh: connect to host savannah.psi.ch port 22: Connection timed out
> svn: Connection closed unexpectedly

sorry, my side network problem.
N.
  298   01 Sep 2006 pohlForumHytec 5331 CAMAC kernel 2.6 driver problem
Grüezi,

I am new to this list.
We are using MIDAS in the Muonic Hydrogen Lamb Shift experiment at PSI. Previously the DAQ was maintained by Paul Knowles. For the upcoming beamtime I took over.

Now I have problems with the kernel driver khyt1331_26 with Midas svn 3315.

I have compiled the driver, and modprobe khyt1331 works.
Then: "cat /proc/khyt1331" gives, with the CAMAC crate switched OFF:

Hytec 5331 card found at address 0xCC40, using interrupt 10
Device not in use
CAMAC crate 0: not responding
CAMAC crate 1: not responding
CAMAC crate 2: not responding
CAMAC crate 3: not responding


When I switch the crate on and do the "cat" again, the computer freezes.
When I switch the crate OFF again, the computer screen turns black and the computer beeps.


Is anybody using the Hytec 5331 PCI CAMAC card plus the Hytec 1331 CAMAC crate controller and can help me?

I would greatly appreciate any help. Otherwise I am lost.

Cheers,

Randolf


More info:
------------------------------------------------------
Using SuSE 9.3 on a P4. Tried HyterThreading on and off.
uname -a:
Linux mpq1p13 2.6.11.4-21.13-smp #1 SMP Mon Jul 17 09:21:59 UTC 2006 i686 i686 i386 GNU/Linux

------------------------------------------------------
This is exactly what I did (my logbook):
> cd $MIDASSYS/drivers/kernel/khyt1331_26
edit kyt1331.c:
replace (line 36):
# include <config/modversions.h>
with
# include <linux/config.h>
now
> make
> make install
Works, but produces irrelevant error:
install: cannot stat `../doc/*.9': No such file or directory
(Some doc stuff missing)
Finish "make install" by hand by typing
> /sbin/depmod

Load the driver and check it is there:
> modprobe khyt1331
> lsmod | grep khyt
gives on my machine:
"khyt1331 13084 0 "

Now try
> cat /proc/khyt1331

Gives on my machine (no CAMAC crate attached)
Hytec 5331 card found at address 0xCC40, using interrupt 10
Device not in use
CAMAC crate 0: not responding
CAMAC crate 1: not responding
CAMAC crate 2: not responding
CAMAC crate 3: not responding

Finally we need the character device with major number 60 ("char-major-60)
called "/dev/camac".
First check that no device with major=60 exitst:
> ls -l /dev | grep "60,"
should not produce any output.
So we create this device by
> mknod /dev/camac c 60 0
And
> ls -l /dev | grep "60,"
results in
crw-r--r-- 1 root root 60, 0 2006-09-01 14:25 camac

(Here start the problems described above. I had the same problems when I tried the "cat" with CAMAC on BEFORE I did the "mknod")

----------------------------------------------------------
Uncommenting all "prink" in ../drivers/kernel/khyt1331_26/khyt1331.c I get the following kernel logs in /var/log/messages:

Sep 1 17:15:55 mpq1p13 kernel: khyt1331: module not supported by Novell, settin
g U taint flag.
Sep 1 17:15:55 mpq1p13 kernel: khyt1331: start initialization
Sep 1 17:15:55 mpq1p13 kernel: khyt1331: Found 5331 card at CC40, irq 10
Sep 1 17:15:55 mpq1p13 kernel: khyt1331: initialization finished
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 0
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 1
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 2
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 3
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 0
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 1
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 2
Sep 1 17:15:59 mpq1p13 kernel: khyt1331: ioctl 3, param 3

And then it dies.
  424   05 Feb 2008 qinzeng pengForumrpc timeout, related to event_size and watch dog? need help
Dear all,

I'm trying to write a simulation code on midas. What I did is just modify the
frontend.c(pp) from experiment samples and made some parameters change on midas.h .

Because my simulation ask for about 4.5MB for each event, so I increase the
MAX_EVENT_SIZE and max_event_size accordingly.

in midas.h :

#define MAX_EVENT_SIZE 0xa00000 //0x400000 /**< maximum event size 4MB*/
#define BANKLIST_MAX 640 //64 /**< max # of banks in event */
#define DEFAULT_RPC_TIMEOUT 60000 //10000
#define WATCHDOG_INTERVAL 5000 //1000
#define DEFAULT_WATCHDOG_TIMEOUT 60000 /**< Watchdog */


in frontend.cpp :

BOOL frontend_call_loop = TRUE;
INT max_event_size = 5 * 1024 * 1024;
INT max_event_size_frag = 2* max_event_size;
INT event_buffer_size = 2 * max_event_size;

EQUIPMENT equipment[] = {

{"WFD_SIMU", /* equipment name */
{1, 0, /* event ID, trigger mask */
"SYSTEM", /* event buffer */
#ifdef USE_INT
EQ_INTERRUPT, /* equipment type */
#else
EQ_POLLED, /* equipment type */
#endif
LAM_SOURCE(0, 0xFFFFFF), /* event source crate 0, all stations */
"MIDAS", /* format */
TRUE, /* enabled */
RO_RUNNING, // | /* read only when running */
// RO_ODB, /* and update ODB */
5000, /* poll for 500ms */
0, /* stop run after this event limit */
0, /* number of sub events */
0, /* don't log history */
"", "", "",},
read_simu_event, /* readout routine */
},
......
}
INT frontend_loop()
{
/* if frontend_call_loop is true, this routine gets called when
the frontend is idle or once between every event */

ss_sleep(100);

return SUCCESS;
}


Compilation OK and running mlogger, odbedit, frontend is OK.
start the run -> no problem ( but there is a long waiting time in frontend if
starting the run. Before the run begins, frontend terminal popping up messages
frequently, say, every 10 seconds. When run starts, frontend terminal hang on
for a couple of minutes before popping up next bunch of messages.)

stop the run -> Problem -> rpc timeout

message from odbedit:

[qzpeng@phy2-dhcp140 simu]$ odbedit -s 10000000
12:54:27 [WFD Simu,INFO] Program WFD Simu on host phy2-dhcp140 started
12:54:37 [Logger,INFO] Program Logger on host phy2-dhcp140 started
[local:simu:S]/>start
Run number [1]: 7
Are the above parameters correct? ([y]/n/q):
Starting run #7
Run #7 started
[local:simu:R]/>stop
[midas.c:9231:rpc_client_call,ERROR] rpc timeout, routine = "rc_transition",
host = "phy2-dhcp140.bu.edu"
Error: Unknown error 504 from client 'WFD Simu' on host phy2-dhcp140.bu.edu
[local:simu:R]/>


runing message from frontend:

[qzpeng@phy2-dhcp140 simu]$ ./frontend
Frontend name : WFD Simu
Event buffer size : 10485760
System max event size : 10485760
User max event size : 5242880
User max frag. size : 10485760
# of events per buffer : 2

Connect to experiment...
OK
Init hardware...
......
  426   06 Feb 2008 qinzeng pengForumrpc timeout, related to event_size and watch dog? need help

Stefan Ritt wrote:
Most likely you changed the maximal event size in midas.h, but you did not re-compile all programs.

Every time I changed midas.h or system header files, I did the re-compile with following procedure:
ipcrm
rm .*.SHM
mcleanup
make clean
make
su
make install

Stefan Ritt wrote:
The maximal event size goes into the size of the shared memory buffer, so all participating programs have to have the same setting, especially the mserver program.

Question here:
How to compile mserver seperately? I think make and make install under midas directory already did the job.


Stefan Ritt wrote:
update to the latest midas version, which is revision 4116

I used latest svn version, so I believe I'm using the latest version 4116.
svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/midas/trunk midas
svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/mxml/trunk mxml

I followed your instructions and did the following:
1)
svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/midas/trunk midas
svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/mxml/trunk mxml
2)
changed two parameters in midas.h
#define MAX_EVENT_SIZE 0xa00000 //0x400000 /**< maximum event size 4MB->10MB*/
#define BANKLIST_MAX 640 //64 /**< max # of banks in event, I need 356 */
3) make
su
make install
I don't know if I need compile mserver seperately.
4) running only two programs:
odbedit -s 10000000
./frontend

And I still got the same problem as before.

[qzpeng@phy2-dhcp140 simu]$ odbedit -s 10000000
[local:simu:S]/>ls
System
Programs
Experiment
Logger
Runinfo
Alarms
[local:simu:S]/>mkdir Equipment
12:44:12 [WFD Simu,INFO] Program WFD Simu on host phy2-dhcp140 started
[local:simu:S]/>start
Run number [1]:
Are the above parameters correct? ([y]/n/q):
Starting run #1
Run #1 started
[local:simu:R]/>stop
[midas.c:9231:rpc_client_call,ERROR] rpc timeout, routine = "rc_transition", host = "phy2-dhcp140.bu.edu"
Error: Unknown error 504 from client 'WFD Simu' on host phy2-dhcp140.bu.edu
[local:simu:R]/>


And I know that the run stopped after a while on frontend, but after the eroor message showed above. If I tried to stop again in odb after a while, it did stopped.
[local:simu:R]/>stop
Run #1 stopped


By the way, thanks for the quick response. I've been working on this for a couple of weeks and I am a newbie.

I also attached my frontend.cpp code and output of make with warning message but comilation completed. Thanks in advance.

In frontend.cpp I only use simulation and don't need any hardware realted issue or function calls but I jsut leave some of them there.