  590   04 Jun 2009 bazinskiBug Reportmhttpd command line experiment specifying
Hi

Not sure how the rest of you specify mhttpd to work with multiple experiments on
one machine, but it would seem not the same as me ;-)

when executing mhttpd with 

mhttpd -e "experimentname" -p "experimentport" -D 

that experiment name is not transfered to transitions as cm_transition never
specifies the experiment in the call to "transition STOP" etc.
the only flag it sends is a -d for debug if selected.

The result is that the stop and start button of the webinterface does not work,
and transitions sit endlessly doing nothing but consuming all the processor,
odbedit works fine though.

Does everyone else use an apache reverse proxy and or explicit experiment choice
in the url ?

As an aside in mhttpd.c in the reply to -? it states 2 -h options the second
should be a -e. line 13378.


Thanks
Sean
  592   05 Jun 2009 bazinskiBug Reportmhttpd command line experiment specifying
Hi

> > Not sure how the rest of you specify mhttpd to work with multiple experiments on
> > one machine, but it would seem not the same as me ;-)
> 
> Please note that there has been a change concerning multiple experiments inside 
> mhttpd. From revision 4346 on, mhttpd can only connect to one single experiment, 
> and the experiment name in the URL (aka ?exp=name) is not supported any more. So if 
> you have several experiments, you start several instances of mhttpd now on 
> different ports.

That i do with : 
mhttpd -p xx -e experiment_name -D

> 
> > that experiment name is not transfered to transitions as cm_transition never
> > specifies the experiment in the call to "transition STOP" etc.
> > the only flag it sends is a -d for debug if selected.
> 
> When connecting to an experiment, any midas client uses the ODB from that 
> experiment so lives in that "namespace". So one client can never call any client 
> from another experiment. So your problem must be something else. Of course there is 
> not parameter "experiment" passed to cm_transition() since the experiment is 
> implicitly defined by the ODB mhttpd is attached to.

Will have to look else where.

> 
> > The result is that the stop and start button of the webinterface does not work,
> > and transitions sit endlessly doing nothing but consuming all the processor,
> > odbedit works fine though.
> 
> I guess you have to do some debugging there. Note that "detached" transitions have 
> been implemented recently by Konstantin, so maybe your problem is related to that. 
> In this case Konstantin should check what's wrong.

cm_transition does a "system(str)" on line 3243 inside the "if(async_flag == DETACH)" of
line 3219, how does an external program know about the state of the originating mhttpd
process ? Surely that str which executes "mtransition ......." should get a -e
specifying the experiment explicitly ? probably a -h as well to be thorough.
The only other way that mtransition.cxx will be able to pull in the experimentname is
from the environment variable in its call to cm_get_environment(....) on its startup.


Ok after some testing .... 
If i start the mhttpd with the environment variable MIDAS_EXPT_NAME set then its happy
as mtransition inherits the environment of mhttpd so cm_get_environment(...) of
mtransition picks up the experiment. Similarly if i insert "-e experimentname" into the
string "str" that is passed in system(str) of line 3243. Then start and stop buttons work. 

Konstantin any comments.

I suppose i can live with starting mhttpd with the environment set before running, but
that kind of negates the command line argument to mhttpd. 

Thanks for the help

Sean
  154   08 Oct 2004 chris pearsonInfoIncreased number of clients in midas.h, important!
> > For one thing, looking at a given midas-using executable, how do I tell what version of midas it has inside?
> 
> Ther is a function cm_get_version() returning the version. As for the executable, all you can do is a
> 
> strings <executable> | grep 1.9

   A lot of programs have a commandline option, such as "--version", where they return the program version number then exit.  As well as the
program version number, the version number of the midas library it's linked with could also be returned (There can be more than one
libmidas.so on a system and this would show which one was currently being linked)

   Something I would find useful would be for the version number to identify precisely which version you have, i.e. not to have different
versions of midas given the same version number.  I've had problems earlier this year due to midas-1.9.3 changing several times between
January and July, while keeping the same number.  I think if "in-between" versions of midas are to be made available, they should contain a
revision number or date or something in the version number to identify them.

> For remote connections (through mserver), there is already a version check. If the minor version differs, you get a warning, if the major
> versions differ (1.>>9<<.4), the client won't start. So at least for remote connection you get a clue.

   Safety measures like these can sometimes get in the way, if you know what you're doing.  So unless there is absolutely no possibility of
success, I think checks such as this one should be overrideable (by a client option).

Chris
  175   24 Nov 2004 chris pearsonInfomidas on 64bit opteron
   Midas, version 1.9.5 of 7th October, was installed, with a few changes, on a
64 bit opteron computer, running linux.  For this processor, as for the alpha
processor, long integers and addresses are 64 bits.  We added a new flag in the
Makefile,

250a251
> ARCH   = $(shell uname -m)
377a379,381
> ifeq ($(ARCH),x86_64)
> OSFLAGS := $(OSFLAGS) -DX86_64
> endif

and extended the alpha-specific definitions, of DWORD and PTYPE, in midas.h to
include this case,

549c549
< #ifdef __alpha
---
> #if defined(__alpha) || defined(X86_64)
598c598
< #ifdef __alpha
---
> #if defined(__alpha) || defined(X86_64)

apart from this, there are a large number of cases where pointers are cast to
integers, without using the PTYPE definition.  These all need to be changed by
hand, although these conversions should probably be removed anyway - in almost
all cases they are unnecessary, as just differences are being calculated.

There were also a number of warnings, which we ignored, where printf format
strings specified long integers, but the argument was not a long integer.  Casts
should probably be added in all cases where the type of the argument can vary
depending on the machine.

A midas analyser was made, which was able to successfully replay some data, but
this was all that was tested.

Chris
  176   25 Nov 2004 chris pearsonForumuse of assert in mhttpd
   We've had mhttpd aborting regularly since upgrading from midas-1.9.3.  This
happens during elog queries, and is due to an elog file that was incorrectly
modified by hand.  The modification to the file occurred 6 months ago.
   el_retrieve(midas.c:15683) now has several assert statements, one of which
aborts the program on reading the bad entry.

   Why is assert used, instead of an error return from the function (if
necessary), and maybe an error message in the log file?  Assert statements are
often removed, using NDEBUG, for normal use.

Chris

   The problem elog entry had one character removed, so end-of-file came before
the end of the message.  This could probably occur without the file being
altered, if the disk containing the elog fills.
  730   02 Nov 2010 chris pearsonInfomhttpd: Extra entries on status page
   A couple of experiments at triumf wanted certain important odb variables
displayed on their status page.  (There was already the possibility to show the
run comment)
   A new folder "/Experiment/Status Items" was created containing links to the
variables of interest, these items are show on the status page, under the run
comment (if any), in 3 columns.

the code from mhttpd.c:show_status_page()
between
   /*---- run comment ----*/
and
   /*---- Equipment list ----*/
is attached
  702   12 Jun 2010 hai quForumcrash on start run
Dear experts,

I use fedora 12 and midas 4680. there is problem to start run when the frontend
application runs fine. 


# odbedit -c start


Starting run #18
[midas.c:8423:rpc_client_connect,ERROR] timeout on receive remote computer info: 
[midas.c:3659:cm_transition,ERROR] cannot connect to client
"feTPCPacketReceiver" on host tpcdaq0, port 36663, status 503
[midas.c:8423:rpc_client_connect,ERROR] timeout on receive remote computer info: 
[midas.c:4880:cm_shutdown,ERROR] Cannot connect to client 'frontend' on host
'hostname', port 36663
[midas.c:4883:cm_shutdown,ERROR] Killing and Deleting client
'feTPCPacketReceiver' pid 24516
[midas.c:3857:cm_transition,ERROR] Could not start a run: cm_transition() status
503, message 'Cannot connect to client 'frontend''
Run #18 start aborted
Error: Cannot connect to client 'frontend'

11:03:42 [Logger,INFO] Deleting previous file "/home/daq/Run/online/run00018.mid"

11:03:42 [Logger,INFO] Client 'feTPCPacketReceiver' on buffer 'SYSMSG' removed
by cm_watchdog because client pid 24516 does not exist

11:03:42 [Logger,ERROR] [system.c:563:ss_shm_close,ERROR]
shmctl(shmid=7274511,IPC_RMID) failed, errno 1 (Operation not permitted)

11:03:42 [ODBEdit,INFO] Run #18 start aborted
==========================================================================

there are several ethernet cards on the host machine. eth0 connect the host
machine to the gateway machine and the front end application listen to eth1 for
the incoming data packets:

eth0      Link encap:Ethernet  HWaddr xx:xx:xx:xx:xx:xx  
          inet addr:10.0.1.1  Bcast:10.0.1.63  Mask:255.255.255.0
          inet6 addr: fe80::f6ce:46ff:fe99:709b/64 Scope:Link
          UP BROADCAST RUNNING MULTICAST  MTU:1500  Metric:1
          RX packets:470870 errors:0 dropped:0 overruns:0 frame:0
          TX packets:515987 errors:0 dropped:0 overruns:0 carrier:0
          collisions:0 txqueuelen:1000 
          RX bytes:345000246 (329.0 MiB)  TX bytes:377269124 (359.7 MiB)
          Interrupt:17 

eth1      Link encap:Ethernet  HWaddr xx:xx:xx:xx:xx:xx   
          inet addr:10.0.1.2  Bcast:10.255.255.255  Mask:255.0.0.0
          inet6 addr: fe80::226:55ff:fed6:56a9/64 Scope:Link
          UP BROADCAST RUNNING MULTICAST  MTU:1500  Metric:1
          RX packets:0 errors:0 dropped:0 overruns:0 frame:0
          TX packets:15 errors:0 dropped:0 overruns:0 carrier:0
          collisions:0 txqueuelen:1000 
          RX bytes:0 (0.0 b)  TX bytes:1836 (1.7 KiB)
          Memory:ec180000-ec1a0000 


thanks for hints
  704   14 Jun 2010 hai quForumcrash on start run
> - does your feTCPPacketReceiver die during the start-of-run? Maybe you do some segfault 
> int he begin-of-run routine. Can you STOP a run?
when start a run, it bring the mtransition process and I guess the server try to talk to the
client, then it fails and the frontend application get killed since not response.

>> When you try to stop your fe from 
> odbedit with  # odbedit -c "shutdown feTCPPacketReceiver"

it gets
[midas.c:8423:rpc_client_connect,ERROR] timeout on receive remote computer info: 
[midas.c:4880:cm_shutdown,ERROR] Cannot connect to client
"feTPCPacketReceiver" on host 'tpcdaq0', port 35865
[midas.c:4883:cm_shutdown,ERROR] Killing and Deleting client
'feTPCPacketReceiver' pid 27250
Client feTPCPacketReceiver not active


what does this error mean? :
11:03:42 [Logger,ERROR] [system.c:563:ss_shm_close,ERROR]
shmctl(shmid=7274511,IPC_RMID) failed, errno 1 (Operation not permitted)


thanks
hai

p.s. that code runs fine on my laptop with ubuntu 9, so that also be possible that somewhere
my configuration not right to cause problem
  2330   08 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
Hi all,

I am having some issues getting the ODBINC command to work within the MIDAS 
sequencer. I am trying to increment one of the ODB values but it is returning a 
mismatch data-type size error (see attached image).

All the ODB variables are MIDAS data-type FLOAT and should all be 32-bit values, 
but for some reason MIDAS is thinking they are 4-bit values. I have tried creating 
new ODB keys of type INT32, UINT32 and DOUBLE but they all give the same error.

If anybody has any suggestions I would really appreciate some help:)

Thanks



 
  2332   09 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> Please post the output of odbedit "ls -l" for /eq/ar.../variables. (you posted the 
> variable name as an image, and I cannot cut-and-paste the odb path!). BTW data size 4 is 
> correct, 4 bytes for INT32/UINT32/FLOAT. For DOUBLE it should be 8. For you it prints 32 
> and this is wrong, we need to see the output of "ls -l".
> K.O.

Hi,

Thanks for getting back to me regarding this. The output of "ls -l" is:

[local:mu3eMSci:S]/>cd Equipment/ArduinoTestStation/Variables
[local:mu3eMSci:S]Variables>ls -l
Key name                        Type    #Val  Size  Last Opn Mode Value
---------------------------------------------------------------------------
_T_                             FLOAT   1     4     1h   0   RWD  20.93
_F_                             FLOAT   1     4     1h   0   RWD  12.8
_P_                             FLOAT   1     4     1h   0   RWD  56
_S_                             FLOAT   1     4     1h   0   RWD  5
_H_                             FLOAT   1     4     60h  0   RWD  44.74
_B_                             FLOAT   1     4     60h  0   RWD  18.54
_A_                             FLOAT   1     4     1h   0   RWD  14.41
_RH_                            FLOAT   1     4     1h   0   RWD  41.81
_AT_                            FLOAT   1     4     1h   0   RWD  20.46
SP                              INT16   1     2     1h   0   RWD  10

Many Thanks
Jago
  2338   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> I tried following script:
> 
> ODBSET /Equipment/ArduinoTestStation/Variables/_S_, 10
> 
> LOOP 10
>   WAIT seconds, 3
>   ODBINC /Equipment/ArduinoTestStation/Variables/_S_
> ENDLOOP
> 
> and it worked as expected. So I conclude the problem must be in your script. Probably a typo in 
> the ODB path pointing to a 32-byte string instead to a 4-byte float.
> 
> Stefan  

Hi Stefan,

Cheers for the reply. I believe the syntax we are using is correct. I have tried copying the script 
you used above and it results in the same error as before. Perhaps something is going wrong 
elsewhere - I will take another look today.

Jago
  2339   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > 
> > [local:mu3eMSci:S]/>cd Equipment/ArduinoTestStation/Variables
> > [local:mu3eMSci:S]Variables>ls -l
> > Key name                        Type    #Val  Size  Last Opn Mode Value
> > ---------------------------------------------------------------------------
> > _T_                             FLOAT   1     4     1h   0   RWD  20.93
> > _F_                             FLOAT   1     4     1h   0   RWD  12.8
> > _P_                             FLOAT   1     4     1h   0   RWD  56
> > _S_                             FLOAT   1     4     1h   0   RWD  5
> > _H_                             FLOAT   1     4     60h  0   RWD  44.74
> > _B_                             FLOAT   1     4     60h  0   RWD  18.54
> > _A_                             FLOAT   1     4     1h   0   RWD  14.41
> > _RH_                            FLOAT   1     4     1h   0   RWD  41.81
> > _AT_                            FLOAT   1     4     1h   0   RWD  20.46
> > SP                              INT16   1     2     1h   0   RWD  10
> > 
> 
> This looks okey, so we still have no explanation for your error. Please post your sequencer 
> script?
> 
> K.O.

Hey, thanks for getting back to me

We are fairly confident the syntax is correct. Having tried the test script posted by Stefan:

> ODBSET /Equipment/ArduinoTestStation/Variables/_S_, 10
> 
> LOOP 10
>   WAIT seconds, 3
>   ODBINC 

The same error is returned:/

We will take another look today.

Jago
  2341   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> Just post here a minimal script which produces the error, so that I can try myself.
> 
> ... and make sure that you have the latest develop version of midas.
> 
> Stefan

Here is the simplest script which produces the error:

WAIT seconds, 3
ODBINC /Equipment/ArduinoTestStation/Variables/_S_

I noticed that "Jacob Thorne"  in the forum had the same issue as us in Novemeber last 
year. Indeed we have not installed any later versions of MIDAS since then so we will 
double check we have the latest version.

Jago
  2343   14 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > I noticed that "Jacob Thorne"  in the forum had the same issue as us in Novemeber last 
> > year. Indeed we have not installed any later versions of MIDAS since then so we will 
> > double check we have the latest version.
> 
> As you see from my reply to Jacob, the bug has been fixed in midas since then, so just 
> update.
> 
> Stefan

We have tried updating using both:

git submodule update --init --recursive

and:

git pull --recurse-submodules

But the error still persists. Is there another way to update which we are missing?

Cheers
Jago
  Draft   15 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > But the error still persists. Is there another way to update which we are missing?
> 
> The bug was definitively fixed in this modification:
> 
> https://bitbucket.org/tmidas/midas/commits/5f33f9f7f21bcaa474455ab72b15abc424bbebf2
> 
> You probably forgot to compile/install correctly after your pull. Of you start "odbedit" and do 
> a "ver" you see which git revision you are currently running. Make sure to get this output:
> 
> MIDAS version:      2.1
> GIT revision:       Fri Feb 11 08:56:02 2022 +0100 - midas-2020-08-a-509-g585faa96 on branch 
> develop
> ODB version:        3
> 
> 
> Stefan

Hey Stefan,

We are running the GIT revision midas-2020-08-a-509-g585faa96:

[local:mu3eMSci:S]/>ver
MIDAS version:      2.1
GIT revision:       Tue Feb 15 16:31:07 2022 +0000 - midas-2020-08-a-521-ge43ea7c5 on branch develop
ODB version:        3


which is still giving the error unfortunately. 
  2346   16 Feb 2022 jago aartsenBug FixODBINC/Sequencer Issue
> > But the error still persists. Is there another way to update which we are missing?
> 
> The bug was definitively fixed in this modification:
> 
> https://bitbucket.org/tmidas/midas/commits/5f33f9f7f21bcaa474455ab72b15abc424bbebf2
> 
> You probably forgot to compile/install correctly after your pull. Of you start "odbedit" and do 
> a "ver" you see which git revision you are currently running. Make sure to get this output:
> 
> MIDAS version:      2.1
> GIT revision:       Fri Feb 11 08:56:02 2022 +0100 - midas-2020-08-a-509-g585faa96 on branch 
> develop
> ODB version:        3
> 
> 
> Stefan

We we're having some problems compiling but have got it sorted now - thanks for your help:)

Jago
  2412   20 Jun 2022 jianrunBug ReportError in "midas/src/mana.cxx"
Dear Midas developers,

When we are running the examples in $MIDASSYS/examples/experiment/, we meet some 
problems when analyzing the results:
1. When we analyze the data using the analyzer: ./analyzer -i run00001.mid -o 
run00001.rz  , we find some bugs: 
"
Root server listening on port 9090...
Running analyzer offline. Stop with "!"
[Analyzer,ERROR] [mana.cxx:1832:bor,ERROR] HBOOK support is not compiled in
[Analyzer,INFO] Set run number 6 in ODB
Load ODB from run 6...OK
run00006.mid:2680  events, 0.00s
"
We think this occurs in the "midas/src/mana.cxx ". How can we solve this?

2. When we analyze the above data, an error also occurs: 
[Analyzer,ERROR] [odb.cxx:847:db_validate_name,ERROR] Invalid name 
"/Analyzer/Tests/Always true/Rate [Hz]" passed to db_create_key_wlocked: should 
not contain "["

We simply fixed that just by replacing the "Rate [Hz]" with "Rate" in the 
test_write in midas/src/mana.cxx 
We are curious whether you can fix the problem permanently in the next version, 
or we are not running the code properly. Thanks!
  1668   28 Aug 2019 lcpForumHistory plot problems for frontend with multiple indicies
hi, 

> > That makes things more 
> > complicated than needed. In the normal FE framework, you can define either several equipment 
> > served by one frontend, or even one equipment linked to several devices. In the MEG experiment 
> > we have one slow control frontend controlling ~100 devices without problem. In the old days 
> there 
> > was a problem that some slow devices could throttle the readout, but since the invention of 
> multi-
> > threaded slow control equipment, each device gets its own thread so they don't block each 
> other.
> 

I agree with Stefan, that it's probably better to run a multi-threaded setup, than individual frontends.

The only place I've ever used the frontend index on startup is when I was testing and building
an eventbuilder.

https://midas.triumf.ca/MidasWiki/index.php/Event_Builder#Example

This might explain, why your history is swapping between frontends, as in the event builder, it gets
reconstructed.

Maybe this helps...

LCP


> Perhaps things have changed in the 10 years since the FGD SC code was written. I can do it 
> differently but doing it that way seemed naturual since around 90% of the frontend code that I
> have see does it that way.
  700   08 Jun 2010 nicholasForumcheck out from svn
do: svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/midas/trunk
midas
shows: ssh: connect to host savannah.psi.ch port 22: Connection timed out
svn: Connection closed unexpectedly
  701   08 Jun 2010 nicholasForumcheck out from svn
> do: svn co svn+ssh://svn@savannah.psi.ch/afs/psi.ch/project/meg/svn/midas/trunk
> midas
> shows: ssh: connect to host savannah.psi.ch port 22: Connection timed out
> svn: Connection closed unexpectedly

sorry, my side network problem.
N.
ELOG V3.1.4-2e1708b5